problem with external Power Supply - 1st Generation

Posted: Tue 31. May 2011, 14:58
by PEAK-Support
If you ever had a very bad looking CAN Scope signal like this,
Bad Screen
Bad Screen
badscreen.jpg (20.42 KiB) Viewed 6908 times
please check if you have connected a GND to the unit on one of the GND Pins (do not forget to switch the GND on in the Device Settings menue - D-Sub GND connection). You also could use the shield of the SUB-D or the GND socket on the back of the Device.
SUB-D
SUB-D
SUB-D.JPG (20.63 KiB) Viewed 6908 times

The power supply with only two pins on the HighVoltage (110-220V) side have no real GND! Only the new power supplies with 3 Pins (Neutral, Phase, Ground) have a real GND.
After you have a GND connection between the PCAN-Diag Device and your CAN Device (CAN-GND), the signal should look like this:
Good Screen
Good Screen
goodscreen.jpg (17.61 KiB) Viewed 6908 times